Python Example code for NI 9229

Yes, it was issue with an 2.7.13. Thanks!

0 Kudos
Message 11 of 12
(1,392 Views)

Hi all,

 

My reading is different than the current probe reading. I see that the possibility could be because the minimum input signal that my device receives is 96 mA (~96 uV) using a sense resistor of 0.1 ohm whereas current probe only measures 30 mA. Is there a possibility that the signal range for NI-9229 might be playing an important role? If so, how do I change the input signal range from +/- 60V to +/-10V in python? Also, what if my one analog input signal range has +/-60V and another +/-100mV? Is there a way to handle each signal range?

0 Kudos
Message 12 of 12