I'm not sure I know what you mean by the "2 counter wide range method." Since you mention an array of values, it sounds like a buffered period measurement -- but that doesn't require a 2nd counter.
When performing a buffered period measurement, the first value's meaning and usefulness depends on whether the operation is triggered.
A) Untriggered operation - the first value is the # of source edges seen from the instant of the software call that "arms" the counter until the first gate edge. This value is meaningless in many applications and you can discard it using the 'Array Subset' function in the array palette.
B) Triggered operation - the first value is the # of source edges seen from the instant of the hardware trigger edge until the fir
st gate edge. This value is much more likely to be meaningful.
You had other questions about the start trigger.
First, for a buffered measurement, there is no retriggering. Only the first trigger edge has any effect on the acquisition. Retriggering is only supported for pulse generation where each trigger edge initiates a single pulse.
Second, while you're
allowed to specify the same external signal as both a start trigger and a gating signal, I don't see much advantage. Maybe you can describe what you would expect as a "correct first value?"
Good luck!
-Kevin P.