Counter/Timer
Externally trigger binned counting
I have a PCI-6221 and am trying to achieve the following.  I've tried searching and found similar questions, but nothing exactly the same, so apologies if I have overlooked a previous answer.
What I would like to do is count total pulses per 100 µs for some fixed period of time, say 100 ms.  The start of the 100 ms total measurement period should be triggered by an external signal which is much slower, e.g. 1Hz  This is as I need to sum multiple mesurement events as follows, where the time bins given are times after the trigger signal and each measurement is triggered by the external signal.
Measurement     0-100µs  100-200µs    200-300µs   etc.
       1                    2                1                0
       2                    1                2                1
       3                    2                1                1
       4                    0                1                0
    total                  5                5                2
If I could get the rolling total for each measurement that would also be fine, I can subtract each time bin from the previous in software.  The previous example using this method would therefore be:
       1                    2                3                3
       2                    1                3                4
       3                    2                3                4
       4                    0                1                1
    total                  5                10              12
I have looked at previous answers such as http://forums.ni.com/ni/board/message?board.id=40&message.id=2056&query.id=1096458#M2056 and whilst I converted that to a finite number of measurements OK, I couldn't work out how to set it to digitally trigger without errors.  I'm using LabView 7.1.
Thanks very much, I hope that this was clear.
Alex
(5,951 Views)
I haven't fully fleshed out this idea, but here it is in a think-out-loud form.
I'm assuming you don't have clocked analog output in your app, and will lean on using an AO task solely for its clock.
Here's the idea:
1. One of your counters is set as a retriggerable single pulse generator.  Let's say it's CTR0.  Use the external ~1Hz signal as its trigger, and set it up to generate a single pulse with a 100 msec high time and a minimal low time.  You'll need a DAQmx property node to make it retriggerable.
2. Configure an AO task to generate at 10 kHz, corresponding to your 100 microsec bins.  Configure it to be pause triggered by the output of CTR0.  Thus, the AO generates samples at 10 kHz, but only while CTR0 output is high (for 100 msec).  You'll have to write a buffer of analog values to generate, but it can just be an array full of 0 voltage, and you don't have to wire that output to anything either.
3. Configure your other counter, CTR1, for period measurement.  This may be tricky to configure with DAQmx property nodes, but ultimately you'll need to treat your pulses of interest as the timebase for the measurement while using the AO sample clock as the signal to be measured.   That sounds backwards, but the effect will be to store the number of your pulses of interest that occur between AO clock edges.  You will need to research "duplicate count prevention" to make sure that 0 values get buffered properly.
I *think* that sort of arrangement could work.  About every second, you get an external trigger.  That trigger causes CTR0 to generate a single pulse with a 100 msec high time.  Your AO task generates a clock pulse every 100 microsec, but only during the 100 msec when CTR0 is high.  Each of those AO clock pulses buffers a value for your CTR1 period measurement.  This should produce a measurement buffer like the first one you listed.
4. (Simpler Alternative) Configure CTR1 for simple counting.  Use DAQmx Timing.vi to specify the use of the AO Sample clock, and specify your external pulses of interest as the signal to measure.  By default, LabVIEW will expect to see those pulses at CTR1's default SOURCE pin.  This arrangement should produce a measurement buffer like the second one you listed, though you may still need to research "duplicate count prevention" to make sure that gets set up right.
5. I can't help but think there's another way to do this without the AO stuff, but nothing's coming to me now.
-Kevin P
Message Edited by Kevin Price on 02-01-2010 02:19 PM
(5,937 Views)
Hi Kevin,
Thank you very much for your response, most useful!  I have implemented both methods you suggested (after a fair bit of muddling through) and both work really nicely.  I need to do a little bit more testing to check I've implemented it properly, but first tests are excellent.  I'll probably go for the first method (the period measurement) as it gives me directly what I want without the need for any subtraction.
Thanks again for the very quick and excellent help,
Alex
(5,916 Views)