Hi:
I am assuming that you are using 2 counters to generate a finite pulse train for 10 seconds and are using this pulse train to gate the other counter. If this is your setup, you should not have a problem in retrieving the correct number of readings.
You can monitor to the backlog and number read by wiring indicators to Read Buffer VI. Please also look at the Count Buffered Edges (DAQ-TIO).vi examples to learn how you can correctly set up a buffered counter acquisition.
To open up the example, click on Help -> Find Examples -> Hardware Input & Output -> Traditional DAQ -> Counter -> NI-TIO.
Regards,
Bharat Sandhu
Applications Engineer
National Instruments