Counter/Timer
PXI-6602 TTL Output Voltage
I have Timer/counter PXI-6602 (for pulse generation, delay and so on) and deliver a T/C out signal with BNC-2121.
(1) Is the Out signal TTL/CMOS?
(2) What is Output impedance?
(3) When I monitor a signal with oscilloscope, the output voltage is 5V with 1 MOhm DC coupling, but less than 2 V with 50 Ohm DC coupling. My thought is that the output voltage should be ~5 V in 50 Ohm coupling.
Can you give me an answer why the voltage is so small? And Can you give me a solution to boost the voltage in 50 Ohm impedance input?
Thank you.
Grace.
(3,589 Views)
Hello Grace,
"(1) Is the Out signal TTL/CMOS?"
Yes, the outputs are TTL Signals.
"(2) What is Output impedance?"
The Output Impedance is 75 Ω (56 Ω from an onboard resistor and 19 Ω from the TIO ASIC).
"(3) When I monitor a signal with oscilloscope, the output voltage is 5V with 1 MOhm DC coupling, but less than 2 V with 50 Ohm DC coupling. My thought is that the output voltage should be ~5 V in 50 Ohm coupling."
Because the output impedance of the 6602 is 75 Ω, a 50 Ω load will sink more current than the 6602 can provide. This causes the output voltage to drop to below 2 V. To prevent this, avoid using a load with such a small impedance. If you must use the 50 Ω coupling, you should put a large resistor (in the kΩ range at least) in series with the load.
I hope this helps.
Sean C.