Counter/Timer
Running 6 counters at the same time
I want to use 6 counters in NI PCIe-6612 at the same time to count pulses from 6 channeltrons, and I found that the counters miss some counts when I use TTL out of a function generator as a test source. For example, it shows ~40 counts in one second when I apply 60 Hz TTL pulse. The fluctuation is also quite large (~30%). When I use only one counter at a time, it reads more accurately, but still ~10% less counts than what it should be.. Is it possible that the computer cannot take the load since it's quite old (about 8 years, Pentium 4 with 1GB RAM).
(6,544 Views)
You need to attach your code. You are hiding everything by using the DAQ Assistant.
(6,514 Views)
How often are you reading the count? I have found that when I use 6612 or 6602 counters in event counting mode with random (or at least asynchronous) events, there is a dead-time after the sample clock pulse is generated to read the count. Therefore, the more often you read the count, the more likely you are to miss counts. I have not seen a problem with using multiple counters but I don't think I've ever used six counters in event counting mode.
All statements are my opinion and worth every cent you paid for them.
Tom Whitaker, CLD
"Give every man thy ear but few thy voice."
Polonius in Hamlet.
(6,447 Views)
It is about once every ~2-3 seconds. I understand the each of them counts simultaneously independent of each other?
(6,440 Views)
The effect I'm talking about requires about a thousand reads to miss an event. Your issue seem to be much worse than this so it is likely not the pulse collision problem I was describing.
Tom Whitaker, CLD
Polonius in Hamlet.
(6,436 Views)