Two pulse trains, one with delayed time

Solved!
Go to solution

Hi,

 

I am trying to get two pulse trains using VI I attached here. I am able two produce two pulse trains but the amplitude of one is very low (0.2V) than expected. Can you please help me how to get same amplitude (5.0V) foe both pulse trains?

 

Regards,

Bharat

0 Kudos
Message 1 of 8
(4,698 Views)
Solution
Accepted by topic author b0giri01

Voltage outputs of counters aren't programmable, so it isn't the program.  The low voltage output is must be connected to a low input impedance circuit.  It's trying to pull more current from the counter than the counter can source, consequently the voltage is pulled way down.

 

On many devices, counter outputs are limited to a small # of mA.  It varies by device, but you can find it in the specs.

 

 

-Kevin P

ALERT! LabVIEW's subscription-only policy came to an end (finally!). Unfortunately, pricing favors the captured and committed over new adopters -- so tread carefully.
Message 2 of 8
(4,666 Views)

Thanks Kevin,

 

Yes you are right but I found that the first pulse train from VI is always with high amplitude and 2nd with low whatever the counters chosen.

I am using PCI-6601.

0 Kudos
Message 3 of 8
(4,659 Views)

What are your counter outputs physically wired to?  Did you change the physical wiring when configuring the task to use different counter #'s? 

 

Each counter has a distinct pin for its output signal.  Are you sure you've identified those pins correctly?

 

 

-Kevin P

ALERT! LabVIEW's subscription-only policy came to an end (finally!). Unfortunately, pricing favors the captured and committed over new adopters -- so tread carefully.
0 Kudos
Message 4 of 8
(4,653 Views)

Hi,

Counter's outputs are connected to oscilloscope directly (No other physical connection) and I didn't make any changes when configuring.

Regarding the pins for output, I checked it again. All are correct.

 

Thanks,

Bharat Giri

0 Kudos
Message 5 of 8
(4,651 Views)

Hi,

I figured out the problem. It's due to the probe I used.

 

Thanks for your help.

 

Bharat

0 Kudos
Message 6 of 8
(4,650 Views)

Just for the sake of future readers, was one probe / scope input at 50 ohms while the other was more like 1 Megaohm?

 

 

-Kevin P

ALERT! LabVIEW's subscription-only policy came to an end (finally!). Unfortunately, pricing favors the captured and committed over new adopters -- so tread carefully.
0 Kudos
Message 7 of 8
(4,645 Views)

Yes the ratio is same.

One of the probe was attenuated by 20 times.

 

Bharat

0 Kudos
Message 8 of 8