Hi all / Karsten,
I found out it was NOT possible to use toggle AND pulse in different terminals.
However I worked a way around it. Later on I saw that it was also possible to trigger acquisition on the 'toggled' signal via the RTSI bus and to use the counter output as my amplifier blanking. I guess one could also use 'pulse mode' and an external flipflop device on another output, with this solution no external electronics were needed.
Finally I used two OUT-counters, clocked with the same timebase and one is triggered on the others up-flank so they are synchronized (otherwise they can have 180' shift!!!!).
I used one "pulse freq." counter to down-divide the 20 MHz clock (master) and the other to count the flanks (slave). Using 'CO Pulse ticks' one can count in the up stage as well as in the low stage, which was convenient for my application (this didn't seem possible with 'CI cnt edges'). In CO pulse ticks: when u set the initial state to 'high' one can only use the 'low time' and 'initial delay', same goes for initial low, only 'high ticks' can be used. I didn't find this in any manual !!!
The blanking and triggering of the acquisition is now HW timed and synchronized with my Wave Form Generation.
Thanks, Greetings,
Paul Knijn