I'm trying to read the frequency of a prox- meter installed as a encoder (reading of a squared-tooth wheel)

 

I'm using Dasylab 11 and MC's USB-1408FS.

 

The problem I'm getting is that the reading obtained by the counter input module is not continuous: when I set a certain speed up, It ''hesitates'' between two value, the real speed (verefied by a stroboscope) being given by an average of these two speed...

 

I've tried setting a faster sampling rate, but dasylab still reads a constant speed as a hesitation between two speed...

 

Any ideas??

 

I've attached a graphic of three speed variation, notice the ''steps'' taken by dasylab..

0 Kudos
Message 1 of 8
(7,845 Views)

Are you using the counter input?

 

Also, what is the frequency you're are measuring (ppr)?

Tom Rizzo
InSyS Corp.
www.insyscorp.com
Your DASYLab integrator
0 Kudos
Message 2 of 8
(7,837 Views)

I am using the counter input, I'm reading 20 ppr at a speed of (maximum) 500 RPM, usually more around 200-300 RPM

0 Kudos
Message 3 of 8
(7,834 Views)

The counter on the USB-1408FS is “software polled”, not read at a hardware pace. That will certainly affect the quality of the incoming signal.

Try reducing the block size to about 1/10th of the sample rate. For example, sample at 100 samples/second, block size of 10.

Measurement Computing (MCC) has free technical support. Visit www.mccdaq.com and click on the "Support" tab for all support options, including DASYLab.
0 Kudos
Message 4 of 8
(7,831 Views)

Also,

 

the spikes you see are probably the mechanical error of the teeth in your wheel.  This is a common effect.  Basically, if you would connect the output to an ossiloscope, you would not see a perfect squarewave but a moving one changing frequency +/- few hertz.

 

The signal shoulod be approximately 166 Hz at 500 rpm so that you could use an analog input as well to measure the speed

Tom Rizzo
InSyS Corp.
www.insyscorp.com
Your DASYLab integrator
0 Kudos
Message 5 of 8
(7,826 Views)

I've change the sample rate and block size to 100 and 10.

 

When I do so for the PC-clock (by un-clicking automatic), I get readings that change from 70 to 80...

 

I have the impression that this particular factor is behind the fact that my readings cannot be as precise as I want...

 

I've tried putting 1000 Hz(sample rate) and 100 as block size, dosen't work also...

 

Thanks for the tips !

Perhaps you've got an idea...

0 Kudos
Message 6 of 8
(7,825 Views)

I will try the analog input solution,

 

Yes, part of my error is definitely mechanical, because the fluke connected to the same input changes a little bit (but not as drastically and clearly separated in to frequencies per speed)....

 

 

0 Kudos
Message 7 of 8
(7,824 Views)

For a good counting module I would suggest

http://www.mccdaq.com/usb-data-acquisition/USB-4300-Series.aspx

 

This unit is designed for counting pulses

Tom Rizzo
InSyS Corp.
www.insyscorp.com
Your DASYLab integrator
0 Kudos