You may get a better answer if you try the forum at www.measurementcomputing.com; they have a DASYLab section, too.
For your device, you need to configure the sample rate for the device using either the Experiment->Experiment Setup->MCCDAQ or the Experiment->Timebase dialogs. Depending on the number of installed devices, you will have one or more tabs for the MCC devices.
You do need a block size that is a multiple of 31, either 4092 or 4991 should work well. Then, use the Data Window module to convert the block size to 4096 (or more or less, as long as it's a power of two.)
Hope this helps,
Measurement Computing (MCC) has free technical support. Visit
www.mccdaq.com and click on the "Support" tab for all support options, including DASYLab.