Hi there!

I'm trying to get triggered scans for measurements. So I have configured in Experiments->hardware setups the trigger as following:
-one scan per trigger
-external rising ttl edge (i also tried all the other possibilities, that doesn't change anything)
When I run dasylab there will be data aquired as soon as the first trigger event starts. but from then on starts a continous aquisition although i only want values when the next trigger signal comes. What configuration do I have to do else?
thx for help!
0 Kudos
Message 1 of 2
(7,936 Views)
The external hardware trigger is only a start trigger. It does not monitor the signal continuously and start/stop based on the signal.

Within DASYLab, use a software trigger, such as the Pre-Post Trigger or the Combi-Trigger Module, in combination with a Relay Module, to monitor the signal and remove the undesired values.

See the Hands On Guide or the online Getting Started Guide for examples of the Trigger modules and how to use them with the Relay.
Measurement Computing (MCC) has free technical support. Visit www.mccdaq.com and click on the "Support" tab for all support options, including DASYLab.