Jon,
Thanks for the reply, i'm using labview 8.2.1 and the latest max + traditional nidaq support.
I got round the problem by setting up the daqpad to run in a handshaking mode, using the c ports and a trigger line and thereby got round any need to rapidly poll line states on the interface. Also ensuring the interface was only initialised once and then left running helped (using advanced io vi's). It now seems to be able to handle transfers at the not particuarly fast 100 ms that i require. Also stability was improved (slightly) by not connecting the daqpad through a usb hub.