We are working on a project utilizing type K thermocouples and the NI- USB 6210 DAQ and when we view the readings in lab view, the readings seem to be opposite from what they should be. For example, when we place the thermocouples over a heater, the readings show that the temperature is actually decreasing, while a seperate infrared thermometer is showing that the temperature has actually increased.
Any help on this issue would be greatly appreciated. Thank you for your time.
Nadim Virani
Aerospace Engineering Student
California Polytechnic State University, San Luis Obispo
07-10-2009 11:02 PM
07-11-2009 03:40 PM
07-12-2009 08:01 AM
Well, I woud suggest having a look at Table 1 (Analog Input Connections) at the following page:
Field Wiring and Noise Considerations for Analog Signals
I suspect that you are using DIFF measurement, so I would recommend using bias resistors when connecting the wires (which is mentioned in the table). Try it out and let us know.
11-16-2009 02:32 AM
Similar problem here, but in my case the offset is much higher, around 10C
I tried to play around with different pullup and bias resistor values, but nothing changed much. I am using:
- SCXI 1102 plus 1303 screw terminal
The resistors I use are the 10Mohm resistors recommended in the manual of 1303 terminal.
Any ideas or things I could try?
11-17-2009 01:37 PM
Regards,
Dan King
11-18-2009 02:13 AM
Hello Dan
After a lot of connections and disconnections I finally made it work:
- As I was using different colouring standards for TC, there was a mismatch of material.
- For the resistors, I finally removed the pull up and set the bias resistor to 10ohm. I do not see however why it does not work with the pull up, because the 1303 manual shows the 10M-pullup & 10-bias configuration as suitable for floating measurements ...
In any case, after these two actions, readings are fine.