I'm currently conducting a project that conducts a burn-in test on a power MOSFET. We are applying 600V betwen the Drain and Source terminals and attempting to monitor the leakage current flowing through the MOSFET in a continuos manner.
The project involves a very high DUT count and therefore we are unable to cater for each DUT to have it's own current measurement channel. We are also unable to switch the cicuit to a shared multi-meter since the requirement is to have the voltage be present throughout the testing.
The existing solution I'm thinking of is to use a sense resistor in series to the circuit. However this will cause the DMM to have a loading effect during the voltage measurement and give erroneous results.
I would really appreciate any feedback on the possible measurement methods I could use in this scenario. My leakage currents range from 100nA to 5mA based on the DUT.
Thank you.
Cheers!
Sanka
10-19-2010 06:02 AM - edited 10-19-2010 06:09 AM
How about using an OP as a current to voltage converter and provide a low impedance voltage to a mux?
OP + to ground
OP - to source
R1 of 1k from OP output to OP -
D1 for protection between OP + and OP - .
+/- 10 to 15 volt supply to power the OP with 100nF Buffer.(maybe 10µF and 2x 10R for additional decoupling)
Without failure the voltage between + and - will be zero -> no current in D. All leakage current will flow through R1 ( plus I_bias) and the OP will provide the voltage (and current) such that your source pin is virtually allways at ground level.
Output voltage will be U_out= - I_measure*R1 (+ U_offset +I_bias*R1).
With R1 = 1k you have a 0.1V to 5V output signal.
NS LMP7721 has a bias in the fA range (but lower supply voltage)