Digital Multimeters (DMMs) and Precision DC Sources
HW-timed vs SW-timed sequence
I ran both of the following examples:
NI-DCPower Hardware-Timed Voltage Sweep
NI-DCPower Software-Timed Voltage Sweep
on my PXIe-4141 SMU with the Source Delay set to 0.
I captured the voltage sweep on an a scope. I was surprised to find out that BOTH examples produced a staircase voltage ramp with the same delay between each step of about 17ms. I expected the HW-timed sequence to be much faster than the SW-timed sequence. (see attached screenshot)
How do you explain this observation?
(4,186 Views)
Hi, AJ3D.  Here's another forum post that I think answers your question:
https://forums.ni.com/t5/PXI/Fastest-update-rate-for-source-operations-in-PXIe-4141/m-p/1923143/high...
And here's a list of attribute defaults for the PXIe-4140/41 http://zone.ni.com/reference/en-XX/help/370736U-01/nidcpowercref/supportedattributes4140_4141/ .  0.016...67 seconds shows up on a couple of different rows: aperture time and source delay.  You minimized only one of those.
Chris
(4,163 Views)
Thanks Chris.
I learn something everyday.
However, I expected the SW-timed sequence to be slower than the HW-timed due to the overhead of CUP-to-SMU communication after each measurement step.
(4,134 Views)
It is slower (and also not deterministic), but not so much slower that it jumps out visually in a scope when each voltage is being held for ~17 ms. The time it takes the computer/driver to reconfigure the device is measured in microseconds.
Marcos Kirsch
Chief Software Engineer
NI Driver Software
(4,125 Views)