Digital Multimeters (DMMs) and Precision DC Sources
PXI 4130 no auxillary output
What part of PXI 4130 SMU possible got damage?
EOS/EMI from AC source can damage the PXI 4130 SMU board?
PXI 4130 SMU board has a protection from the high frequency noise?
(4,364 Views)
Hey Jvmaster
How are you using this device? Are you sourcing and staying within the specifications for the device below?
PXI-4130
http://www.ni.com/pdf/manuals/374746e.pdf
Also, could you attach any warnings or errors that you are receiving that makes you think something is damaged on your device.
Gerald M.
NI
(4,332 Views)
Hi Gerald, the system is working at the production line, it is only due for calibration, so need to pull-out the SMU board from the PXI Chassis and bring to central calibration office. during calibration when the calibration software request to install the auxiliary power supply for verification, then declared no aux power out supply
(4,275 Views)
Hey Jvmaster,
So, you took the chassis out, took it your calibration team, and they determined that something was wrong with the system. Did they specify any further what the issue was? Is the AUX power out not reaching levels they expect? Are the following the calibration procedure for this module?
CALIBRATION PROCEDURE
NI PXI-4130
http://www.ni.com/pdf/manuals/374753e.pdf
Also, if they do believe the module is broken we do offer repair services. I will attach some more information on that below.
Repair and Replacement Services
http://www.ni.com/services/repair/
National Instruments Hardware Repair or Calibration
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P6lYSAS&l=en-US
Gerald M.
NI
(4,247 Views)