Hello ERT_LSU,
There are two possibilities why the setup is not working for you. I assume you have the SCXI-1127 connected with an AUX cable to the DMM 4060. This means that the trigger between them is sent using that same AUX cable. Therefore you should select the Trigger Input as the Read Connector of Module 4. This ensures that the switch works properly.
Another possibility is that the problem is with the setup instead. I recommend measuring one of electrodes straight with the DMM to make sure that the connections are correct and that you can indeed read the right values. If this works then try with only two channels witch the switch. Make sure that the Scan List of the program is updated properly to "ob0!sc1!md4!ch0->com0;ob0!sc1!md4!ch1->com0;" if your connections are to channels 0 and 1.
I have tested the example and I was able to measure 9 and 10 volts by switching two different channels. Make sure your devices work properly. Try using the Test Panels in MAX to check your measurements before getting into such a large setup. This might narrow down the problem.
Let me know how it goes.
Regards,
Tica T
Applications Engineer
04-03-2006 03:32 PM
Hello Tica T,
Do you mean Rear connector of Module 4 instead of Read connector? The trigger input is already set to the Read connector of Module 4. I also followed the commands you wrote to me. But it's still giving small values. I'm sending you a screenshot of the GUI of the program from my computer. Can you send a screenshot from your computer showing all the configurations and how it's getting 9 V measurement? Thank you very much for your response.
Sincerely,
ERT_LSU
04-03-2006 03:32 PM - edited 04-03-2006 03:32 PM
Message Edited by ERT_LSU on 04-03-2006 03:36 PM
04-03-2006 04:56 PM - edited 04-03-2006 04:56 PM
Hi ERT_LSU,
In order to get the SCXI-1127 configured and working correctly you will need two types of connections. The banana jack connections will allow the DMM to connect to the output out0 of the 1331 and therefore read the correct voltage depending on the switch state. You also need a backplane adapter for controlling the SCXI module. This could be an SCXI-1359 connected on the back of the switch and an AUX cable going from the AUX in of the 1359 to the AUX of the 4060 DMM. You could also have some other type of controlling device for the SCXI chassis.
The parameters you use in the VI are going to be dependent of the system setup. First I want to verify your configuration of the SCXI, how do you have the SCXI chassis connected to the computer?
The way the chassis is connected will dictate whether the chassis is recognized as a DAQmx chassis or a Traditional DAQ chassis.
If you are using the SCXI-1359 configuration then the chassis will show up under Traditional DAQ devices in MAX and the configuration will probably be something similar to what it is shown here:
Have you tried connecting only a single electrode to the DMM to check that the DMM is measuring the right value?
The values I have are very similar, only the Instrument Descriptor and the Chassis Descriptor are different and they come from the MAX configuration.
Let me know how these measurements go.
Regards,
Tica
Applications Engineer
Message Edited by Tica T on 04-03-2006 04:56 PM
04-04-2006 03:10 AM
04-05-2006 05:32 PM
04-07-2006 04:34 PM
Hi Partha,
What I meant by ‘measuring voltage from a single electrode’ is the following: instead of measuring voltage using the switch, try measuring the electrodes you have connected to a single channel by connecting them directly to the DMM to make sure that it is reading the correct value (up to 12V as you specified).
Once we know that the measurements from a single channel are correct then we can worry about setting up the switch.
From what you describe on your connections though, everything seams to be connected properly. The SCXI-1359 is the connector you are using on the back.
The DMM is actually capable of measuring resistance by direct connection. I am attaching the Datasheet and here is a link to the product help that explains how to make the connections to measure resistance.
Hope this helps. Let me know if you are able to measure the right values.
Regards,
Tica T
Applications Engineer
National Instruments
04-25-2006 09:25 AM
04-28-2006 08:50 AM
Hi Partha,
At this point the best solution is to call in for support to speak to one of our Applications Engineer in the Modular Instruments team. This will help us troubleshoot the problem step by step and if it turns out to be a broken switch we can also do an RMA on it.
You can request a service request at the following link: http://sine.ni.com/apps/utf8/niae_asc.main
Good luck,
Tica