Dynamic Signal Acquisition
Help calibrating the PCI-4474 Board
Spoke with a rep from NI. They said in order to externally calibrate the PCI-4474 board, I should use the same procedure as the 4472. My question is this: The steps are:
1. connect the calibrator voltage to all input channels.
2. Set the voltage to 5.00V
3. Check the date (using nidsacal.dll)
4. Call Calibrate_DSA (I am assumiing I can use DSA Calibrate.vi. This is where I run into problems. I am suppose to set "setofCalConst" to ND_USER_EEPROM_AREA. The calibrate.vi does not have this option. Am I missing something. Then step 5 is to
5. Call copyCalConst to move the info stored in the user area of the EEPROM to the factory area of the EEPROM. This step I have no problem with because it is called from nidsacal.dll
(6,543 Views)
Hi Kimberly,
Actually, you do not need to specify the ND_USER_EEPROM_AREA. The prototype in the DLL is Calibrate_DSA (deviceNumber, operation, refVoltage) and you only need to specify the device, the operation (internal or external calibration), and give the reference voltage. These should be the same three inputs you have available from the DSA Calibrate VI in LabVIEW. Simply specify the device, external calibration, and give the reference voltage and you're set.
Best Regards,
Justin Britten
(6,543 Views)