Hello Mr Hinrichs,
Could you please give me more detailed information’s, about where you found this data’s about THD and SFDR. I had a closer look to the specifications of the USB-9211, but I couldn’t find any information in the user manual USB-9211/9211A User Guide and Specifications Are you sure, you named the right devices in your post?
I’m waiting for your response, so we can go further with this issue
Kind regards
Ulrich Möschel
Application engineer
07-26-2006 02:49 AM
01-21-2018
04:48 PM
- last edited on
07-15-2024
07:12 PM
by
Content Cleaner
Total Harmonic Distortion is the sum of distortions over a frequency range.
Spurious Free Dynamic Range is the maximum spur across a frequency range and is not related to harmonics as harmonics are a multiple of the frequency.
More can be found here:
https://www.ni.com/docs/en-US/bundle/ni-scope/page/spurious-free-dynamic-range.html
THD Should always be greater than SFDR as THD is a sum of harmonics whereas SFDR relates to one specifically which is the maximum across a spectrum.
01-22-2018 09:50 AM
Yes, harmonics are often excluded from SFDR measurement, and that appears to be the case in the published specifications for the 9234.
With a 9234 input, at a test level of 3.5 V pk (-3.1 dB ref 5 V), I consistently measure SFDR to be 112 dB or better.
01-22-2018 10:04 AM
THD Should always be greater than SFDR as THD is a sum of harmonics whereas SFDR relates to one specifically which is the maximum across a spectrum.
Generally, there is no set relationship between THD and SFDR. If you have a case where you are including harmonics in SFDR and the max spur is a harmonic of the test frequency, then |THD| <= SFDR.
01-22-2018 01:05 PM
sorry, must be usb-9233
I refer to the 9234 in my responses because that is NI's current recommended 4 Ch, 24-bit, c-Series device.