Dynamic Signal Acquisition
Will NI-9222 or NI-9223 work with PCB IEPE (ICP) sensor?
Greetings,
I have an application where i need to use PCB ICP pressure sensors (Model: 102B) that have the following specs
Measurement Range: 5kpsi
Sensitivity: 1.0mV/psi (hence output signal +/-5VDC)
Excitation Voltage: 20 to 30VDC
Constant Current Excitation: 2 to 20mA
Output Bias Voltage: 8 to 14VDC
We need cRIO platform because of rugged environment. Now i know that cRIO9234 would be good choice, however there are two showstoppers.
1- I must use exernal excitation source from PCB for higher current (7 to 10mA) because of 170ft cable length and to overcome cable capacitance. For this reason i can not use cRIO9234 because it has fixed in 2mA current excitation
2- Secondly my frequency of interest is 20 to 35KHz so again 9234 will not cut it.
Question is can i use 9222 (500KS/s/ch) or 9223 (1MS/s/ch) modules for this application or will i miss something by not using 9234?
Also having SAR (Successive Approximation) ADC versus delta sigam ADC makes problems in readings? NI doesnt clearly say 9222 o 9223 are meant for ICP pressure or accelerometers, therefore i am a little concerned.
Kindly share your input experiences.
(11,220 Views)
actually the module you should consider is the 9232
https://www.ni.com/en-us/shop/model/ni-9232.html
It has 41kHz bandwidth when sampling at 102.4kSA/sec.  It is 24bits, and offers 4ma of power per channel.
you will see with a 5V P-P signal from the sensor, 4ma gives you about 35kHz bandwidth at just under 200 feet of cabling.  Lower P-P signals go farther.
If you desire 10ma (I do not think you need it), you can use a low cost PXI chassis with the 4462 module which offers 10ma of excitation.
Hope this helps.
Preston Johnson
Solutions Manager, Industrial IoT: Condition Monitoring and Predictive Analytics
cbt
512 431 2371
preston.johnson@cbtechinc
(11,214 Views)
Thanks. preston can you post a correct article link?
Also 3 module module will not work considering the number of channels we need and 8 slot cRIO chassis we have. Will 9222 and 9223 will not work?
(11,207 Views)
Sorry about that - here is the article
https://www.ni.com/en/support/documentation/supplemental/08/using-long-cables-with-iepe-sensors.html
Preston Johnson
cbt
512 431 2371
(11,201 Views)
Can I use PCB pressure sensor 113B26 with NI 9234 module?it's measurement range is 500 psi and excitation needed is 2to 20 mA. Please guide.....
(5,473 Views)