Hi,
I have bought the NI PXI4472B. I have 2 identical sensors which have a frequency range of 0.03-50Hz. With these sensors, I measure the velocity of the ground. Unfortunately, I have a good coherence only from 1 to 50Hz.
The problem is that the resolution of my ADC is not good enough : even if I have 24 bits of resolution, the full scale is -10V/+10V, and the signals of my sensors are in the maximum range of -50mV/50mV. So I would like to know if it is possible to adjust ADC gains to have a full scale of the maximum amplitude of my signal for example.
Please help me,
ben
Sanaa T.
National Instruments France
>> http://www.nidays.fr/images/081110_ban_nidays09_468X60.gif
06-30-2005 03:18 PM
07-04-2005 12:47 PM
07-19-2005 03:48 AM
... that is the fine thing about 24 bit ADC: You typically do not get a higher sensitivity, you just get rid of choosing the input range 😞
The limiting factor often is the analog design: cross-talk, noise, etc. This really makes a high quality (expensive) measurement system. Today, you get 24 bit converters, giving you a theoretical dynamic of nearly 150 dB. But you will not get that in practice, because you cannot do that easily in analog. And you cannot do it in anti-aliasing, too. So, what you typically get at the specs is something about 120 dB, for real world conditions it is often worse.
If you go down 120 dB, starting at 10 V, you end up at 10 muV. If you have an old, traditional, simple low dynamics 16 bit system giving a 10 mV input range (and many did so) and 80 dB dynamic range (even at 10 mV!), you end up with 1 muV, i.e. a 20 dB higher sensitivity for low-level signals. And you can be sure that there is visable aliasing coming up in your measurement data, etc...
These traditional systems have only these drawbacks:
So, as in many likewise situations: Remember the old instruments and have a close look on performance degradations of "new, simple" solutions (likewise reasoning applies for high precision, low signal measurements comparing charge type accelerometers with ICP / IEPE type ones)!
just my 2 cent,
schi
07-19-2005
08:25 AM
- last edited on
02-18-2025
09:59 AM
by
Content Cleaner
There are also devices which have both a 24-bit ADC and gain which provides even higher sensitivity. An example of this is the PXI-4462.
07-19-2005 08:53 AM
So, the device is really fine (according to the spec, I did not test one by now), but confirms my earlier remark.
schi
07-19-2005 09:39 AM - edited 07-19-2005 09:39 AM
Message Edited by benbol on 07-19-2005 09:42 AM
Message Edited by benbol on 07-19-2005 09:42 AM
07-19-2005 10:00 AM
07-19-2005 01:37 PM
Thanks for your answer. I am working with accelerometers, so I need IEPE supply, and so I have to use AC coupling, if not my DAQ input is saturated. I don't understand the reason, because the output impedance of my sensors are less than 10 ohm, do you know why? Do you know if I can find a DAQ, when used on AC coupling, which have a high pass filter whith a cutoff frequency less than 0.1Hz? The other solution is to use an external supply for my accelerometers, so I will be able to use DC coupling, do you think it's possible?
I am also working with sensors which measure velocity and which have an external supply. So we don't need IEPE supply in this case and I use DC coupling. I don't understand in this case why the DAQ input is not saturated, do you know why? Can't I use a digital high pass filter (for instance in Labview) instead of an analog high pass filter? Because I'm afraid of the noise I can add.
For the time constants of the filters, you are true : it's approximatively 10 seconds when using IEPE powered sensors with the PXI4472B. So I wait before recording my data. Is it because of high pass filter (with a cutoff frequency of 0.5Hz)? Or because of the antialiasing filter?
In general, do you think the PXI4472B is suited for low frequencies mesurements (0.1Hz to 100Hz for ground displacement measurement with nanometer scale) if I use DC coupling? I don't know if +-10V full scale with 24 bits of resolution is enough.
We want to have a DAQ with several channels simultaneously sampled. Eight channels will be perfect, if not we can synchronize multiple devices. It will be also good to have a sampling rate which will be variable from around 50Hz to 1000Hz or more, which is not the case of the PXI4472B.We would like to have a very good resolution, because we measure very small velocity or acceleration(50mV is the maximum amplitude) For the moment, we are working with Labview (DAQmx and Sound and Vibration Toolkit) with Windows 2000 professionnel, but I am open minded to every better solution.
Thanks a lot for your answer,