1. Select the Fluke Calibrator connected to the system.
2. (Optional) Select the Fluke Self Test and Serial Number tab and run the vi.
3. Select the Verify DC Voltage Accuracy tab.
3a. Input the NI 9210 mini-TC Channels you wish to verify, if data should be logged, the file path data should be logged to, and run the vi.
4. Select the Adjust DC Voltage Acurracy tab.
4a. Input the NI 9210 Module you wish to adjust, the NI 9210 mini-TC channels you wish to adjust, and run the vi.
5. Select the Verify DC Voltage Accuracy tab.
5a. Input the NI 9210 mini-TC Channels you wish to verify, if data should be logged, the file path data should be logged to, and run the vi.
See the NI 9210 Mini-TC Calibration Procedure for further information.