FieldPoint Family
4-20 mA current source not reading properly on AI-110
Thanks for your help. I will get a loop isolator and give it a try. It will probably take a week or more to get it but I'll try to remember to post the outcome.
Thanks again.
Message 11 of 14
(4,107 Views)
It would be worth a quick check with a meter to verify my theory on the chassis of the instrument being tied to the common of the FP is correct.
Message 12 of 14
(4,105 Views)
With the controller and FP AC cords unplugged, I got 54 ohms resistance between the controller chasis and the com terminal of the FP.
Message 13 of 14
(4,102 Views)
Can you try to measure a known voltage source with your AI-110 to ensure it is not damaged?  It also would be a good idea to compare measurements in MAX to those in a VI.  Try measuring a battery or something to see if that works properly.  If it does then connections are most likely to blame.  Can you provide a schematic of your tachometer and FieldPoint connector block?
Eric A.
Distributed I/O Product Support Engineer
Message 14 of 14
(4,037 Views)