FP-TC-120 Temp. readings are erratic

I can only get stable readings if my TC's are similar lengths. As soon as I have TC's with different lengths the longest TC readings go crazy. Is there any way to obtain stable TC readings with different lengths of TC's. I am ready to junk this system so any alternative equipment suggestions would be welcome - preferably not NI equipment
0 Kudos
Message 1 of 3
(3,222 Views)
From your description, I would suspect that your thermocouples are in an environment that is highly electrically noisy, such as being near a motor. The FP-TC-120 module does not have channel to channel isolation, and can only reject several volts of common mode voltage between each channel. Since the readings become erratic when one thermocouple is longer than the others, this is a very good indication that external common mode noise is being induced on the thermocouple leads. As long as the leads are the same length, the same voltage gets induced on all the thermocouples and the module is able to reject the voltage. In order to remedy your situation, I would offer the following suggestions:
1) Make sure the FP-TC-120 common is free to float, to allow it to
better reject common mode noise. This is done by making sure that you have not wired the C or Com terminals of the FP-TC-120 to the C or Com terminals of any other FieldPoint module.
2) Use shielded thermocouple wires. Connect the shield of the thermocouple to the C terminal of the FP-TC-120.
3) Route the thermocouples away from the electrical noise source.
4) Do not ground the ends of the thermocouples. If they need to be touching a metal object, use an insulator and thermal conduction grease.
5) Route the thermocouple cables through grounded metal conduits.

Regards,
Aaron
LabVIEW Champion, CLA, CPI
Message 2 of 3
(3,222 Views)
Thanks for your comments. I believe you are right about the source of the problem being electrical noise since the erratic readings stop as soon as I shut the burners.

I am using shielded lead tc wires up to the furnace ( my furnace is 50'x12'x12') and high temperature tc wires in the furnace. Customer specifications require that TC's are spot welded to the parts being heat treated.

I also notice that the erratic behavior starts around 600 deg. F. Any other suggestions?

best regards
anzio