FieldPoint Family
compensation for FP-RTD-122 rtd
Hi
Iam using FP-RTD-122 with a 3 wire PT100 RTD to measure the TEMP in a Vacuum Chamber. I had got my RTD Calibrated & find that there is a diff of +/- 3 deg w.r.t to my FP-RTD-122 modules.The same RTD when connected to Eurotherm 2408 Temp Controller gives a correct reading. I suspect that there could be some problem in the FP modules, Is there any VI where i can directly write compensation to the modules. Can u plz help me resolve this issue. Cheers Suresh.R
(3,396 Views)
Have you tried unplugging and reconnecting the I/O module while the network module is running? We saw several I/O modules that would show a +4C offset if powered up with the rack but no offset if hot plugged.
Matt
(3,395 Views)
If you are seeing this type of offset, it is most likely the result of ESD damage to the FP-RTD-122 module. The good news that you can workaround this problem by hot plugging your modules into your bank as Matthew has mentioned.
Bill B
(3,395 Views)