Hi Gretchen-
How are you determining these readings? Are you reading in MAX or LabVIEW (or other means)? Are you sure that your configurations for thermocouple type are correct in MAX? I would also suggest that you try your thermocouple readings with a different thermocouple to make sure the problem lies in the FieldPoint hardware or connections and not with the thermocouple itself.
The CJC sensor is just a thermistor embedded in the terminal base, so it doesn't need to really be calibrated. If you would like to verify its specifications you can use these steps:
To verify the connections of your thermocouples you should consult the FP-TC-120 and cFP-TC-120 Operating Instructions. You should connect each of your signals to the appropriate IN(+) and IN(-) pins on the connector block.
If these suggestions do not help please give us a better idea of exactly how you're connecting and monitoring your thermocouples and CJC channels.
Thanks-