Thanks for contacting National Instruments with your issue, we'll try our best to resolve it for you as quickly and efficiently as possible.
While you can specify any vertical range you want with niScope Configure Vertical.vi, from the actual A/D converter perspective, if you want to utilize the full 14-bits of resolution available, then you can't go any lower than the specified 0.2 V. The A/D is scalable only to the specified vertical ranges of 0.2, 0.4, 1, 2, 4 and 10 V whereby the entire 14-bits of resolution will be spanned across the entire voltage range.
So in your circumstance, if you specify a vertical range of 0.002 V, you'll onlly be using 1% of the 14-bits, so as long as that accuracy is acceptable for your application you should be fine.
Best of luck with your project.
Sincerely,
Minh Tran
Applications Engineering