High-Speed Digitizers
Can any high speed digitizers do continous acquisition?
Message 1 of 8
(7,180 Views)
Currently (1/30/01), none of the high speed digitizers (NI-5102, NI-5911 and NI-5112) can do continous acquisition.
Message 2 of 8
(7,180 Views)
As of NI-Scope 2.0 (current version is 2.1.4), you can now perform continuous acquisition on the NI 5112, NI 5620, and the NI 5621.
Message 3 of 8
(7,179 Views)
Hi Josh,
Can you point me at a refernce for that?
I though the scope cards could collect faster than the PCI bus could save, so I am suprised to hear this.
Ben
Message 4 of 8
(7,179 Views)
You are correct! Some digitizers can acquire data faster than the PCI bus can transfer it, so currently the highest sampling rate you can aquire continuously with is around 30MS/s (using the 5112 and depending on your PC of course). We are working on our driver to get the number closer to the specifications of the PCI bus. Page 5-7 of the NI-Scope Software Users Manual will give you more information on continuous data acquisition with NI Digitizers.
Message 5 of 8
(7,180 Views)
Cool!
Keep pushing the envelope.
Ben
Message 6 of 8
(7,179 Views)
Hi josh,
How fast can a PXI 5112 transfer data to harddrive in the controller card.
Message 7 of 8
(7,179 Views)
Josh doesn't the PCI transfer speed depend on whther the card is operating as a Master or a Slave. In master modes the speeds should be much higher than 30MB/s.
Message 8 of 8
(7,177 Views)