High-Speed Digitizers
Can you specify the allowable values for vertical range and offset for the NI 5112 PCI Digitizer?
The NI 5112 manual indicates that the vertical range can be set from +/- 25mV to +/-25V in 10% steps. Can you explain exactly what this means? (For example, could you give me a series of allowable values.)
Also, what are the allowable values for voltage offset?
The reason for my question is: I am specifying the vertical configuration using the niScope_ConfigureVertical() function. The observed raw (8 bit) sampling does not appear to change at all for small changes in voltage range and offset. I would like to know where the hardware breakpoints are, so that I can apply correct scaling from bits to voltage.
In addition, I have tried to use the niScope_CheckAt
tributeViReal64() function to "check" a value before applying it to NISCOPE_ATTR_VERTICAL_RANGE or NISCOPE_ATTR_VERTICAL_OFFSET, but it fails in the linking step in Microsoft Dev. Studio. (niScope.lib & niscope_32.lib do not seem to contain any of the CheckAttribute functions that are listed in the NI-Scope function reference.)
Thank you for your response.
(5,377 Views)
There is a nice reference at:
http://digital.ni.com/public.nsf/3efedde4322fef19862567740067f3cc/38f18d418e2433ed8625690e006bbe94?OpenDocument
which explains that allowable voltage ranges can be expressed by the equation:
next_range_value = range_value * 1.1 starting at +/- 25 mV.
When a range is specified that does not match this series, the device driver rounds UP to the next value.
(5,377 Views)