NI-Scope bug: range property don't give correct value with 5922 5900 combination

I already post this bug in another thread , but with another topic, but since no CAR is provided and the update to 2012 (NI-Scope 3.9.4) didn't fix it,  I bring it up here again.

 

When using NI-Scope with a 5922 and a 5900 preamp the vertical range property is not correct. I use this property to calculate and inform the operator about the the actual ratio of used range choosen by the driver and the actual input signal range.

 

A test-vi is attached and show the error also with simulated devices.  (The actual range should be 40V )

 

NI-Scope range bug.png

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 1 of 8
(10,163 Views)

Hi

 

Why do you not use a differential measurement which allows for  40V peak-to-peak (+/- 20V)?

Is the problem that you do not have enough ports to conduct differential measurements instead of Common Mode measurements?

Simply configure for differential and connect the ground to the negative input terminal. As far as the 160V range showing up, that's definitely some sort of bug.

 

0 Kudos
Message 2 of 8
(10,143 Views)

Hi,

 

If you look in the "detailed help" for the "NI-SCOPE Initialize With Options" VI you will see:

 

You can also use the option string to attach an accessory such as the NI 5900 to your digitizer session to allow the seamless use of the accessory:

Option String: DriverSetup = Accessory:Dev1

Refer to the example niScope EX External Amplifier for more information.

 

It could be that your formatting is slightly wrong (spaces in the string), could you test it and and try to match the string exactly (With spaces)?

0 Kudos
Message 3 of 8
(10,127 Views)

I mean  🙂

                                               string.png

0 Kudos
Message 4 of 8
(10,126 Views)

 

Any updates? It's been a while and nothing have change....

0 Kudos
Message 5 of 8
(10,114 Views)

@hsm78 wrote:

Hi

 

Why do you not use a differential measurement which allows for  40V peak-to-peak (+/- 20V)?

Is the problem that you do not have enough ports to conduct differential measurements instead of Common Mode measurements?

Simply configure for differential and connect the ground to the negative input terminal. As far as the 160V range showing up, that's definitely some sort of bug.

 


The input limit to GND for each single input of the 5900 is +-10V (actually 11V) , so every potential to that is bigger than that will drive the input buffer into its limits BTDT 😞

 

If you drive the inputs with 10V amplitude and 180° phase you have 40Vpp .. see link in my first post.

 

The main reason for this thread is, that I want a CAR for the range bug...

 


And no missspelling in the init with options 😉  Hey, I extra made a demo and explained how to reproduce that error even without having the cards....

 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 6 of 8
(10,107 Views)

Still would be happy about a NI confirmation and a CAR

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 7 of 8
(10,060 Views)

Thanks for the bug report.  I've filed a CAR to our R&D department for NI-SCOPE.  (CAR ID#: 376074)

Systems Engineer
SISU