05-30-2013 09:35 AM
Hello,
I have a similar problem like in the post of NI-scope PCI-5124 with 2 triggers. http://forums.ni.com/t5/High-Speed-Digitizers/NI-scope-PCI-5124-with-2-triggers/m-p/1132555/highligh...
I tried to use this code with my PCI-5105 NI-scope card, but I wasn't able to get it to run using an analog trigger to start the recording.
What do I have to pay attention to make my input signals acceptable as start trigger source?
Thank you,
Philipp
05-31-2013 10:39 AM
Hi Philip,
You cannot use an analog trigger as a start trigger. If you look at the NI High-Speed Digitizers Help file you can see what types are supported for kind of trigger. You can get to this help file in LabVIEW or by navigating to the Windows Start Menu and selecting All Programs >> National Instruments >> NI-SCOPE >> Documentation >> NI High Speed Digitizers Help file. To get to the page talking about supported trigger types within the help file expand Fundamentals and then select Triggering. Hopefully this helps!