I try to reset PCI-5922 from NI-MAX and obtain error code 80040383:
"There was a problem resetting the device."
Then I try self-test from NI-MAX and obtain Status Code: -223431
"The self test failed.
Internal Software Error occurred in Scopes software. Please contact National Instruments Support."
I tried th following:
1. restart computer
2. uninstall and then install NI-Scope 15.1.1
3. change PCI slot
(5,773 Views)
I've found old issues regarding the error code 80040383 and were fixed re-starting the NI Device Loader service, take a look at the following KB to troubleshoot this:
Why Are NI Device Loader and NI Configuration Manager Services Not Starting and Giving an Error -88705?
http://digital.ni.com/public.nsf/allkb/0819391CA91B3EB38625731D00024649
In regard of the other error code I've found that typically when this error shows up the device needs to be RMA'ed, but let's give it a try starting the service I pointed above to see if this helps, if not I recommend you to open a new Service Request to troubleshoot the error 223431.
Luis G.
(5,754 Views)
Hi Luis
I tried but It didn't help.
I'll open a Service Request
Thank you
Yonathan
(5,733 Views)
Accepted by topic author yonathan
After some experimentation it turns out that the problem was in the computer. When installed in another computer, the PCI-5922 worked properly.
Now it seems obvious to me: when troubleshooting
1.) install the board in a different computer
2.) install another identical board (if you have one) in the suspiciuous computer
Yonathan
(5,610 Views)