Hi everyone,
I'm testing the NI-PXI 5922 to check if it is suitable for my application.
In particular, I want to measure a pulsed waveform ranging from -1.7 V to 0 V and I would like to use the minimum allowed range (± 1 V) since I’m actually interested just in the higher part of the signal (around 0 V).
I developed a LabVIEW software that performs a triggered acquisition so I’m sure that, when the trigger occurs, the signal is well inside the ± 1 V range (no pre-trigger samples are acquired); attached, you find a sketch of the acquisition configuration.
The result is that I get the following warning:
“Code: 1074115590
ADC for one or more channels was overloaded. Your data might be invalid.”
Of course, if I repeat the same acquisition with the ± 5 V range, there is no warning anymore.
The waveforms acquired with the two different ranges, attached in this message, are pretty different for about the first 50 us. With the ± 1 V range there is an overshoot that I think is due to the recovery of the board from the overload condition.
Do you know if there is any way of disabling the board’s channel (with all its circuitry) until the trigger occurs in order to avoid the overload condition (and the resulting recovery)?
11-12-2013 03:57 AM
ps: I don't think that 50 us of recovery for a High-Speed Digitizer is acceptable...
11-13-2013 07:41 AM
You can not do that. The range 1V, means measure voltage between -1V to 1V. Not +-1V. Whenever you measure voltage higher than that you get an overload. What you could do (not recommended by me) is measure differentially to -1V privided by any powersource. Than 0V would be measured as +1V. But as stated, that is not recommended, mainly you should have a resistor from AI- to ground as in figure 5 on the following webpage:
How To Measure Voltage http://www.ni.com/white-paper/7113/en/
That would mean a constant current from your powersource to ground. That resistor is build in mainly to avoid ghosting effects, and protect the hardware.