"In terms of connecting your thermistor to the USB-6210, you can simply take a differential measurement across two analog input channels (for example pins 15 and 16). Then, you can configure a DAQmx task to automatically convert that voltage into a temperature by creating a DAQmx temperature task. "
That's not true, at least with the combination of USB-6210 and LabView 2009. You will get an error for a two lead thermistor. Any guesses why? No internal voltage or current source? What's the proper way of using a two lead thermistor with the 6210?
07-23-2010 01:20 PM
You are talking about apples and oranges. Your initial question was about cjc and the answer was correct - you do not need cjc with a thermistor. You do need excitation (much different than cjc). If you specify internal excitation, then you will get an error since the DAQ card has no excitation source. You can specify external excitation and design a known current source or use something like the SCXI-1120 which does have an excitation source.