Trigger position is invalid??? pxi-5102 digitizer

Im currently trying to set up 2 channels to read at 8MS/s a peice for 1 second. Im getting the error " Trigger position is invalid." altho.. I dont even see a trigger position option from where I am gettin the error. Im getting the error at the Scope Horz Config VI. I have a picture of the setup Im useing, please le me know if im doing somthing wrong, thank you veyr much, im very puzzled over this.


-Mark
0 Kudos
Message 1 of 2
(5,896 Views)
The 5102 cannot have more pretrigger points than it has onboard buffer memory - 663,000 bytes. By default, the trigger position is 50%, or halfway through the buffer. Taking 8 million points/channel on both channels allows you 331,500 points/channel or a maximum trigger position of about 4%.

Set the trigger position using the reference position input of niScope Configure Horizontal Timing.vi. If you want to always be safe, use zero.
0 Kudos