When do offset and gain change for niScope on a PXI-5122?

Solved!
Go to solution

In my application, I am using the niScope Fetch (2D I16) VI to fetch from a multi-record, multichannel acquisition on a PXI-5122 digitizer.  In trying to optimize my code, this led me to this question.

 

I'm taking 32 successive acquisitions and summing (averaging) the waveforms together.  Based on the advice in Scaling and Normalization of Binary Data, I first apply the wfminfo.gain parameter to the I16 data, sum that, then apply the offset.  In matlab-like pseudocode for my Labview code, it's something like:

 

sum = zeros(1,1000)
offsetsum = 0
for i = 1:32
    [wfminfo, samples_i16] = niscope_fetch_i16(...);
    samples_float = wfminfo.gain .* float(samples_i16[0,:]);
    sum = sum + samples_float
    offsetsum = offsetsum + wfminfo.offset
end
sum = sum + ones(1,1000) .* offsetsum

 

 

So my question is whether it is safe to assume that the wfminfo gain (and offset) parameters will remain constant for the duration of an acquisition session (from the time that niScope Initiate Acquisition is called to the time that acquisition completes)?   The above reference on the normalization coefficients indicates that these coefficients are a function of vertical resolution and calibration settings.  That would allow me to move the gain multiplications outside the acquisition loop and perform integer arithmetic.  Is this a safe assumption?

 

 

 

0 Kudos
Message 1 of 4
(6,551 Views)

I guess this topic should probably be moved to the High Speed Digitizers board.

0 Kudos
Message 2 of 4
(6,541 Views)
Solution
Accepted by topic author gregoryng

Hi Gregory,

 

We tested this out here, and you should be able to pull those gain and offset parameters before the while loop is entered, and perform the arithmetic later outside of the loop.  To do this, you can place a niScope Get Wfm Coefficients VI before the while loop, set it to Get Normalization Coefficients, and route those elsewhere, as they will stay constant during the acquisition.

0 Kudos
Message 3 of 4
(6,507 Views)

Thanks for looking into this for me.  I can see that it definitely changes when the card is reset or self-calibrated, and seems to be stable even between calls to Initiate Acquisition.  I used the "niScope EX Binary Acquisition" vi.

0 Kudos
Message 4 of 4
(6,474 Views)