compatibility ni 5911/5122

I made an application with Labview 6.1 for the PXI-5122 card installed on the PXI-1002 device. I´m trying to run this application on an other computer with the 5911 card. And I have 2 problems:
1)For the 5122 the resource name is PXI1Slot2 (see attached file screenshot 5122) but I can´t find the right resource name of the 5911 (see attached file screenshot 5911). Which resource name name do I have to enter in my program ?
2)As soon as I start the .exe file I have several error messages because some subroutines are missing:
- can´t load NI visual scope.vi
- labviewerror.vi
- niscope close.vi
How can I proceed to avoid these errors and have my program running with the 5911 ?
I have windows 2000 on my both pcs.
Download All
0 Kudos
Message 1 of 2
(5,735 Views)
The PXI-5122 is a DAQmx device, so it's resource name is configurable in MAX (PXISlot2). However, all other National Instruments digitizers at this time are Traditional DAQ devices (including the 5911) and their resource name is "DAQ::n" where "n" is the device number found for the device in MAX. Also, please consult the getting started guide for more information.