niScope Multi Fetch.vi:7390001<ERR>Maximum time exceeded before the operation completed. Status Code: -200284

2018-08-30_11-20-17.png

 

Based on this article (https://natinst.force.com/KnowledgeArticleDetails?id=kA00Z0000019LabSAE)

TimeOut >= SamplesToRead/SampleRate

 

  1. what does SamplesToRead correspond to in association with NI High Speed Digitizers? The following 2 parameters could mean the same thing:
    1. Record length (for example a 1000 sample record)
    2. number of samples (always defaulted to -1 to acquire all samples available)
  2. My current sampling rate = 512M and RecordLength = 2.1M. Given these parameters, duration for a single shot acquisition should be 4.10156 ms. So TimeOut should be > 4.1 ms. I have added 10ms additional overhead to this as shown in figure below.
    1. LabVIEW_2018-08-30_11-29-41.png
  3. The acquisition was working well (was able to fetch 500 records and write to disk) before I closed the VI and re-opened to run another sequence. And it throws this error.

Any advise is appreciated.

 

 

NI System Configuration:
- NI PXIe-1071, 4-Slot 3U PXI Express Chassis , 1 GB/Slot throughput, Part Number: 781368-01
- NI PXIe-PCIe8381,x8 Gen2 MXI-Express for PXI Express Interface,3m, Part Number: 782522-01
- PXIe-5160 PXI Oscilloscope, 500 MHz, 10 bits, 2.5 GS/s, 2 Channels, 64 MB, Part Number: 782621-01
- Astronics PXIe-1209 2-Channel, 100 MHz PXI Pulse Generator, Part Number: 785033-01