06-30-2014 12:19 PM
We can only read 2 digits (before the decimal point) on an MKS deviceNet mass flow controller using Labview 2013. Does anyone know a solution to increase the resolution on this device? We should be able to read 3 digits past the decimal point.
07-01-2014 04:49 AM
Hi,
Is the device a 3rd party device?
According to our engineer, we have no control on how the 3rd party device returns data. Could you clarify a little more about your question?
07-02-2014 12:46 AM - edited 07-02-2014 12:47 AM
Hi David,
Could you also specify the type of mass flow controller device?
According to our engineer, we had the APIs to set the attributes of DeviceNet device, but the specific type of the device is required. Attached is the document of some device type. You could turn to page 75 to see there the S-Analog Sensor Object Attributes::FlowDataType could be Get/Set.
07-02-2014 03:55 PM
The MFC is an MKS GE-series. Our problem is that when we write and read the MFC through Labview-> DeviceNet, we only get integer values. We need 2 or 3 decimal places. Is there a problem with DeviceNet, or the MFC or do you think the problem is with our Labview code?
07-07-2014 01:53 AM
Hi,
Could you kindly paste your VI(s) that utilizes the DeviceNet APIs?
07-22-2014 06:23 PM
check class x31. att 3 and 4 Presume you should be set for real and either sccm or liters. Do same for setpt