Hello,
The 34401A is not longer available in Canada. We need to change and use the 34461A as a replacement and we are having trouble with the peak-to-peak measurements (We are using LabVIEW).
We use the peak-to-peak option while reading a communication line in order to get the min and max values. The code:
Visa Write
34410A
CONF:VOLT:AC
VOLT:AC:PEAK:STAT ON
INIT
FETC?
Visa Read
Visa Write
FETC:VOLT:AC:PTP?
Visa Read
This code no longer works with the 34461A. Reading the AC will only give the RMS values. We dont really care how to get the min-max of the communication channel, as long as it is in a respectable time (10s at most...). Reading the signal with DC pretty much stays around the 0V. The communication signal is ±100mv. We verify the min-max range to be around 200mV.
Thanks
03-16-2015 05:17 PM
03-16-2015 05:40 PM
The problem is that the array of values are all around 0mV when reading the signal with DC. Either single values in a loop or using the Read Multiple points. Which works fine on AC or frequency reads. But the AC reads are always RMS. I want real values, similar to the acquisition with a DAQ. Or just the highest/lowest value over a period of time.
03-16-2015 05:44 PM
03-17-2015 12:54 AM
There is an application note on the keysight site about the differences between the 34401A and the 34461A.
One point says that "the 34641A also suppports overlapped command processing between INIT and measurement completion"
It looks for me that your code depends on the 34401A non overlapped command processing and you have to add some syncronisation with *WAI or *OPC? there.
03-17-2015 12:54 AM
There is an application note on the keysight site about the differences between the 34401A and the 34461A.
One point says that "the 34641A also suppports overlapped command processing between INIT and measurement completion"
It looks for me that your code depends on the 34401A non overlapped command processing and you have to add some syncronisation with *WAI or *OPC? there.
03-17-2015 04:23 PM
if that was the only problem 😛
Many SCPI lines are now giving "Undefined Header" when copy pasted from the Truevolt-DMM-Operating-and-Service-Guide-ENG-2015-2
03-17-2015 04:47 PM
03-18-2015 12:47 PM
These are a few commands which I noted that always give undefined header:
VOLT:DC:SEC "PTP"
VOLT:AC:PEAK:STAT ON
and others I haven't noted.
Those commands aren't the only lines that I write through VISA. I send other strings before and its at that moment that I get the undefined header error. This includes trying with *WAI and *OSC
Example form the guide:
[SENSe:]DATA2:CLEar[:IMMediate] Clears the latest result(s) of the secondary measurement. This is typically used to clear the accumulated history of the peak-to-peak function ([SENSe:]VOLT:[DC:]SECondary "PTPeak" or [SENSe:]CURRent:[DC:]SECondary "PTPeak"). Parameter Typical Return (none) See example below Configure DC voltage measurements using the 10 V range with 1 mV resolution. Select peak-to-peak as the secondary measurement. Make and retrieve measurements. CONF:VOLT:DC 10,0.001 VOLT:DC:SEC "PTP" READ?;DATA2? Typical Response: -1.84694879E-03;-2.23115575E-03,+5.98403324E-04,+2.82955909E-03 (Data returned in this order: DCV reading, Min Peak, Max Peak, Peak-to-Peak) Later, clear peak-to-peak secondary measurement data. Make and retrieve measurements. DATA2:CLE READ?;DATA2? Typical Response: +1.29177114E-03;+6.07221096E-04,+3.43677658E-03,+2.82955548E-03 (Data returned in this order: DCV reading, Min Peak, Max Peak, Peak-to-Peak)
CONF:VOLT:DC 10,0.001 (this works)
VOLT:DC:SEC "PTP" (undefined header)