Acquire waveform from Tektronix 3052C scope with some interval
Hi All,
I am using tektronix 3052C oscillosope. I use averaged method of acquisition. the process I am following is as below
Initialize scope
configure scope parameters (average -64)
call read waveform.vi
get the waveform and measure Vpp and verify.
close the Interface
this thing works fine with no problem.
Now I want to put read wavform vi in the while loop so that I can run it again if Vpp verification fails.
I tried to put only read waveform.vi in the loop, I started getting time out error from fetchwaveform.vi in the 2nd trial.
If I put the whole thing in the while loop, it does not have any problem, but it consumes some time as I have to run initialize and configure Vis again.
If some one can throw some light on myself, that would be really great.....
Anil Punnam
CLD
LV 2012, TestStand 4.2..........
(3,607 Views)
What acquisition mode are you using? Single or Continuous?
(3,592 Views)
Hi
the visa init and close should anyway stay outside the loop, but some of the scope configuration is probably needed to trigger the scope or maybe you are reading too fast for the second waveform
greetings from the Netherlands
(3,589 Views)
@Dennis
I am using continuous mode of acquisition. I believe that is the one required if we want averaging.
@Albert
I kept break point between 2 reads when I was testing this. So there is no problem with the time it seems. I will try to move some of the confguration outside and see the result
Anil Punnam
CLD
(3,585 Views)