Agilent MSO8104a resolution
I am trying to acquire thousands of waveforms from an Agilent MSO8104a using the driver examples provided. Currently I am acquiring continuous waveform data from the scope over lan, but the data I am getting is limited to 8 bit resolution which simply is not enough to capture the waveforms I am collecting. The waveforms range in amplitude from ~-10mV to ~-5V and I have used a DC offset to utilise the maximum range of the ADC, however when I capture a wavefrom on the scope manually I get far superior resolution in both time and voltage.
Is there a way I can increse the resoltion that LabVIEW is retrieving from the scope?
(3,852 Views)
Accepted by topic author jonelmer
Careful, scope displays can play tricks on you.  If you dig through the specs of the scope enough you will find that the ADC in the scope is 8-bit.  You can get 12-bit equivalent with averaging turned on, but that is only good for periodic waveforms.
(3,844 Views)
It seems strange then that the scope appears to save waveforms at a higher resolution, when the waveforms I am working with are non-uniform.
(3,841 Views)