I try to connect two MFC5850s from Brooks with a NI PCI8431/2 RS-485 card.
Those devices are in two wires mode (without RTS or CTS control).
On the RS-485 side, I have connected RX+ & TX+ together and RX- & TX-.
The + common wire is plug on the 'Common A(+)' pin and the - common wire is plug on the 'Common B(-)' on each MFCs.
I have verified the RS-485 parameters (Baud rate, stop bit, parity...) all is ok.
I have configured the COM port in "2-wires mode auto".
I have try to establish communication with Smart DDE from Brooks, without results.
I cannot communicate with the MFCs.
Is somebody have an idea to help me ?
Thank you very much for your help.
09-14-2010 09:08 AM
I reply myself,
Use the NI-SERIAL 3.6 driver and never the 3.5 version !
It correct a bug that don't permit to change the wiring configuration of the RS-485 card (that do be set on 2-wires auto mode).
09-15-2010 08:20 AM
NI-Serial 3.6 did correct a bug where the default wire-mode from Measurement & Automation Explorer would not be used. It was still possible to change the wire mode if it were changed programmatically through your application. I agree that upgrading the driver is the best solution for this problem.