Can't access output2 on afg320
Malosh:
I'm happy you're having some success.
So, here's my take:
a)  Your MAX settings are good.  No need to mess with them or re-install IVI Compliance Package.
b)  Your specific driver works.  No need to re-install it.
c)  I think the issue is that your class driver thinks it is communicating with a single-channel Tektronix function generator.  In your screen shots you have a False wired into the Initialize VI for Identification Query.   This means the initialize routine cannot determine if you are communicating with a single channel model or a dual channel model.   The specific driver is probably set up to default to a single channel model.   So, Channel 2 is not valid.   To fix this, enable the Identification Query when connected with a live instrument.   Alternatively, you can use Initialize with Options and specify the model you are using.  You'll need to read the help to determine the exact syntax for setting this.   I believe this can also be set in MAX too.
The simpliest solution is to enable ID Query...
Message 11 of 12
(1,126 Views)
I changed the ID query to TRUE and woohoo!Smiley Very Happy
Cheers Noel!
Now thats a new-years present!!!
Thanks a lot,
  Malosh
Message 12 of 12
(1,116 Views)