Cannot see GPIB device in MAX after connecting another device to the same port!
I'm controlling a SUSS-prober PA200A using Labview to move around the chuck (for optical measurements). But now when i connect the HP-UX machine to the same prober's GPIB port (on top of this one), to get some electrical measurements, I can't see the GPIB device anymore on the Labview PC.
Basically i need to sync both my electrical and optical measurements but unfortunately have to use two different machines (PC and HP-UX) seperately.
Hope this helps!
PS: Can't use CIC or parallel polling as the prober gets invisible to MAX after parallel connection with HP-UX machine 😞
Jerry
(3,493 Views)
the most likely cause for this is having the same address set on both devices.  Connect the each one at a time and see if they both respond in MAX (and look at the address).
It is aslo possible that one of the devices has a hardware fault in the GPIB interface.  the same test will show you if the devices both work correctly.
(3,491 Views)
And you certainly cannot connect two system controllers at the same time. If you have the GPIB controllers in the pc and hp-ux set to be cic and a system controller, change one. You will need to do quite a bit of modification to code for synchronization if you need them to both to be a controller at different times.