HI all,
I have connected an external light meter to my PC via a RS232->USB converter and have written code to obtain results from the meter.
This works about 75% of the time.
The code performs the following.
1) Open VISA session.
2) Reads in 32 bytes using #$0D as terminator char
3) Closes VISA session.
The meter continually sends data.  Start byte is #$02, end byte is #$0D, all other bytes are ascii.
Problem is, my code does not always 'grab' the bytes.
Can someone show me a working example of code that succeeds in capturing streaming data ?
Bob.
(4,323 Views)
Open VISA serial port once at the beginning of your program and set the termination character to x0D and the timeout say twice the sending rate of your meter.
In a (while) loop read the data with one serial read and ignore the first string read since you will sometimes start your open VISA during a data transmission. If the meter always send the same amount of data, you can use the bytes read value to check your data.
After you terminate your loop (stop button)  close the VISA port
Greetings from Germany
Henrik
LV since v3.1
(4,310 Views)
Hi Henrik,
Thanks for the information.
I now have 100% working code.
Thanks !!  I owe you a weissbier.  🙂
Br,
Bob.
(4,301 Views)
A good rating will help until I get my liquid bread Smiley Very Happy
Henrik
LV since v3.1