Continuous Oscilloscope Acquisition

Hello,

 

My oscilloscope (Agilent DSO-X-2004A) can sample at 1GS/s, but can only window 100kS. I need that sampling rate, but for more than 100us (longest window for 1GS/s and 100kS). Can I acquire one 100kS set and begin recording the next 100kS set immediately after? In that way, I could patch together a data set spanning my desired time length. Otherwise, can I collect continuously from the scope?

 

I connect the scope to my computer via USB (could use ethernet too) and I have LabVIEW 2012 SP1 32-bit

 

Thanks for the help,

John Oxaal

0 Kudos
Message 1 of 3
(3,802 Views)
0 Kudos
Message 2 of 3
(3,798 Views)

When you calculate 100kS on usb2 (8bit) it takes 800kbit and some overhead at 480Mbit per sec so to transfer 1Mbit is 1/480 sec which is 0.5ms

And your scope is probably doing only that. 

So it looks good but way more than the 100us you would need.

 

sorry

greetings from the Netherlands