the function niDMM_Fetch sporadically returns the warning code 1073356801 (0x3FFA2001 "Over range.") with a measurement of 105 to 140 MOhm when using Auto Range. However the subsequent check with niDMM_IsOverRange returns VI_FALSE. As I understood
Can anyone explain this or is this a bug?
Thanks in advance and best regards,
Michael
Solved! Go to Solution.
06-21-2022 09:55 AM - edited 06-21-2022 09:56 AM
So, for all those who are looking for the answer, NI R&D has confirmed this behavior as a bug (no. 2035720).
It occurs every time a resistance above 100 MOhm is measured with the PXI-4082 and Autorange enabled.
Best regards,