E3631A wont responde
I try for example the simplest command, �IDN?�, in HyperTerminal or a similar program to get feedback from HP power supply E3631A. But I don�t get any response. After a while I get �Input buffer overflow� on the instrument display, so I seem to have contact. I have the correct baud rate and settings, cause if I change the baud rate on my computer, I get �RS-232 framing error�.
Neither do I get the �OUTP ON� to work.
What do I do wrong??
Regards Magnus
(4,048 Views)
Sounds like a bad baudrate setting. Can you verify that all of your settings match what the hardware expects? You'll need to check:
Baudrate
Data bits
Stop bits
Parity
Flow control
(4,048 Views)
My settings:
E3631A
Baud rate - 9600
Parity - None
Databits - 8
Start bits - 1 (fixed on instrument)
Stop bits - 2 (fixed on instrument)
Flow control - no settings
COMPUTER
Baud rate - 9600
Parity - None
Databits - 8
Start bits - 1 (fixed)
Stop bits - 2
Flow control - none (I have tried HW as well)
I have double checked!
(4,048 Views)
Hmmm - unfortunately if the settings match and you are having problems with hyperterm, then you may need to contact HP to debug the meter.
(4,048 Views)
Problem solved!
After opening the instrument and resoldering the RS232 connector everything seems to be working just fine. It must have been a loose contact!
(4,048 Views)