Error command with NI Interactive Control
Dear All,
I got a problem when I use NI Interactive Control(NIIC).
I can successfully connect my PC and Instrument.
After I write the command "*OPC?", I could get the correct return data.
(correct return data = Instrument Name)
But when I try another command, it will show error.
I write command: CALCulate:Data:SNP:PORTs:SAVE "1,2,3,4","C:\User\Instrument\Downloads\test11302.s4p"
and push the "write button"
After the above action, I got the wrong message.
Error 1: Return Data form show "Write operation error" and error code "VISA: Hex 0xBFFF00A6"
Error 2: It changed my command text(Red Square)
"User\Instrument" was changed to "UserInstrument" (Blue line)
"Downloads" was changed to "rownloads" (Green line)
This part is the most confusing for me.
PC: Windows 10
Software: NI Interactive Control
Hardware: USB-GPIB-HS
Instrument: PNA N5227B
(1,916 Views)
The string uses a ' \' code display, to use a '\' you need to use '\\'
This means '\n' is used to denote a new line character, '\s' is used to indicate a space constant and so on.
Santhosh
Soliton Technologies
(1,896 Views)
Also note that the
"CALCulate:Data:SNP:PORTs:SAVE "1,2,3,4","C:\User\Instrument\Downloads\test11302.s4p"
refers to a directory on the instrument, not your PC hard drive.  So the directory must exist, or you get an error.
You can also shorten the command to ..
CALC:DATA:SNP:PORT:SAVE "1,2,3,4", "C:\User\Instrument\Downloads\test11302.s4p"
Craig
(1,864 Views)